AD7190 Issues [Copy link]

I recently worked on AD7190, and the sampling results when no-load are roughly as follows:

Is this correct? I don't understand, please help me. Thank you!

This post is from stm32/stm8

Latest reply

If you are not sure whether the ADC is working properly, you can short IN+ IN- to the ground and see the result. If the ADC problem is ruled out, check the analog front-end circuit. Or you can convert the current data into signed decimal data and collect multiple data. Use EXCEL and other software to analyze and see if it is consistent with your expectations.  Details Published on 2019-12-29 18:25
Personal signature为江山踏坏了乌骓马,为社稷拉断了宝雕弓。
 

3180

Posts

0

Resources
2
 


Sometimes it happens again, also empty.

This post is from stm32/stm8
 
Personal signature为江山踏坏了乌骓马,为社稷拉断了宝雕弓。
 

9702

Posts

24

Resources
3
 
Do not leave the input pin floating as the measurement will not make sense
This post is from stm32/stm8

Comments

I mean the sensor is connected, but no force is applied.  Details Published on 2019-12-29 18:16
 
 

3180

Posts

0

Resources
4
 
littleshrimp posted on 2019-12-29 15:18 Don't leave the input pin floating. It is meaningless to measure it.

I mean the sensor is connected, but no force is applied.

This post is from stm32/stm8

Comments

If you are not sure whether the ADC is working properly, you can short IN+ IN- to ground and check the result. After eliminating the ADC problem, check the analog front-end circuit. Or you can convert the current data into signed decimal data and collect multiple data. Use EXCEL and other software to analyze and see if it is consistent with your expectations.  Details Published on 2019-12-29 18:25
 
Personal signature为江山踏坏了乌骓马,为社稷拉断了宝雕弓。
 
 

9702

Posts

24

Resources
5
 
chenbingjy posted on 2019-12-29 18:16 I mean the sensor is connected, but no force is applied.
If you are not sure whether the ADC is working properly, you can short IN+ IN- to the ground and see the result. If the ADC problem is ruled out, check the analog front-end circuit. Or you can convert the current data into signed decimal data and collect multiple data. Use EXCEL and other software to analyze and see if it is consistent with your expectations.
This post is from stm32/stm8
 
 
 

3180

Posts

0

Resources
6
 

Thank you! I'll try it

This post is from stm32/stm8
 
Personal signature为江山踏坏了乌骓马,为社稷拉断了宝雕弓。
 
 

Just looking around
Find a datasheet?

EEWorld Datasheet Technical Support

EEWorld
subscription
account

EEWorld
service
account

Automotive
development
circle

Copyright © 2005-2024 EEWORLD.com.cn, Inc. All rights reserved 京B2-20211791 京ICP备10001474号-1 电信业务审批[2006]字第258号函 京公网安备 11010802033920号
快速回复 返回顶部 Return list