Help: Puyuan DS5102CA waveform measurement problem [Copy link]

I bought a Puyuan DS5102CA oscilloscope ten years ago. Suddenly, no matter what waveform I measure, as long as there is a signal, it is a 50HZ sine wave. However, the 1KHZ 1V signal that comes with the test instrument has no problem. I don't know what the problem is. Maybe it's broken. There is no problem measuring 1K signal. Does anyone know? Thank you, please advise.
This post is from Analog electronics

Latest reply

Yes  Details Published on 2019-5-6 13:23

374

Posts

3

Resources
2
 
The shielding line of the oscilloscope test pen is broken.
This post is from Analog electronics

Comments

Thanks, it is indeed! I changed a test pen and it is fine. Has anyone encountered this problem?  Details Published on 2019-5-6 10:30
 
 

2w

Posts

0

Resources
3
 
It is speculated that the shielding layer of the connecting wire behind the probe (a cable with a shielding layer on the outside) is broken.
This post is from Analog electronics
 
 
 
 

2865

Posts

4

Resources
4
 
wenyangzeng posted on 2019-5-6 10:21 The shielding line of the oscilloscope test pen is broken.
Thank you, it is indeed! I changed a test pen and it is fine. Have you encountered this problem?
This post is from Analog electronics

Comments

Yes  Details Published on 2019-5-6 13:23
 
 
 
 

2865

Posts

4

Resources
5
 
It’s really hard to spot from the outside!
This post is from Analog electronics
 
 
 
 

374

Posts

3

Resources
6
 
bigbat posted on 2019-5-6 10:30 Thank you, it is indeed! I changed a test pen and it was fine. Have you encountered this problem?
Yes
This post is from Analog electronics
 
 
 
 