1993 views|13 replies

3232

Posts

0

Resources
The OP
 

ADC Resolution and Resolution [Copy link]

 

ADC resolution and resolution
1. How to understand the effective resolution of ADC and the gain error of PGA? 2.
What does the resolution of 60000d mean?
3. Are the accuracy and resolution of ADC the same concept ?

This post is from Test/Measurement

Latest reply

CS1237 can float up with a breath of output. Normally it can be used as a 16-bit or 17-bit ADC. If it is higher, it needs to be shielded and temperature-controlled.   Details Published on 2024-5-16 17:33
 
 

6830

Posts

11

Resources
2
 

ADC resolution and resolution
1. How to understand the effective resolution of ADC and the gain error of PGA? 2.
What does the resolution of 60000d mean?
3. Are the accuracy and resolution of ADC the same concept ?

Are you asking about accuracy and resolution?

Precision refers to the error, which is the deviation of the measured value from the actual value.

Resolution refers to the number of bits of the ADC, which determines the scale of the ADC. For example, a 10-bit ADC is 2 to the power of 10, and so on. This is another netizen's answer, I hope it will be useful to you:

The accuracy is related to the entire system. For example, how many volts does your system measure for 1V?


If only the temperature drift of the reference voltage is considered


A system with a reference voltage of 2.048V and a 12-bit ADC measures a 1V DC input. Theoretically, it should be: m1=2.048/4096=2000. Then the actual voltage obtained based on this sampling value is: 2000*2.048/4096=1V

If the reference voltage changes to 2.5v, then the actual measured value is: 1638.4. Considering that the sampled value is an integer, the actual value may be 1638. Because the reference voltage is the "reference voltage" of the system (a bit confusing), then we think it is accurate, that is, we think it has not changed, so it should be 2.048v. In this way, the voltage calculated based on this sampled value is: 1638*2.048/4096=0.819v, and the error is: 18.1%, which is actually (2.5-2.048)/2.5=18.08% (because the sampled value is an integer, so this is slightly different from 18.1%)


From here we can see that although the minimum scale of the ADC has not changed, the accuracy is very poor. This can be corresponded to the commonly used ruler: the minimum scale of the ruler has not changed, but due to thermal expansion and contraction, the ruler has become longer (or shorter), so the size of the measured object is also wrong, but this ruler can still distinguish which of the two different objects is longer and which is shorter according to the minimum scale, but the absolute value is no longer correct. This is the relationship between the resolution and accuracy of the ADC


From the above example, we can roughly get the method of selecting the ADC reference source. Without considering the errors of other parts of the system, the measurement error = reference error (in fact, since other parts of the entire system are not ideal, the measurement error > reference error)

Assuming a 12-bit ADC, the resolution is 1/4096. If we want to achieve an accuracy of 1/4096, that is, for an actual 1v input, we hope that the measured result is between 0.999755859375 and 1.000244140625.
In this way, the error of the reference power supply is <1/4096. Assuming that our system operating range is 30℃, the corresponding temperature drift is: (1/4096)/30*1e6=8.13ppm/℃
————————————————
Copyright statement: This article is an original article by CSDN blogger "varding", which follows the CC 4.0 BY-SA copyright agreement. Please attach the original source link and this statement when reprinting.
Original link: https://blog.csdn.net/varding/article/details/19135249

This post is from Test/Measurement

Comments

I've seen this post, it's not very popular  Details Published on 2023-4-12 18:40
I've seen this post, it's not very popular  Details Published on 2023-4-12 18:38
I've seen this post, it's not very popular  Details Published on 2023-4-12 18:37
 
 
 

6830

Posts

11

Resources
3
 
Also, please check after posting to see if the title is correct.
This post is from Test/Measurement

Comments

It should be resolution and accuracy. Sorry.  Details Published on 2023-4-12 18:36
 
 
 

3232

Posts

0

Resources
4
 
lugl4313820 posted on 2023-4-10 07:16 Also, please check after posting to see if the title is correct.

It should be resolution and accuracy. Sorry.

This post is from Test/Measurement
 
 
 

3232

Posts

0

Resources
5
 
lugl4313820 Published on 2023-4-10 07:14 ADC resolution and resolution 1. How to understand the effective resolution of ADC and the gain error of PGA? 2. What does the resolution of 60000d mean? 3. The resolution of ADC...

I've seen this post, it's not very popular

This post is from Test/Measurement
 
 
 

3232

Posts

0

Resources
6
 
lugl4313820 Published on 2023-4-10 07:14 ADC resolution and resolution 1. How to understand the effective resolution of ADC and the gain error of PGA? 2. What does the resolution of 60000d mean? 3. The resolution of ADC...

However, due to thermal expansion and contraction, the entire ruler becomes longer (or shorter), so the size of the measured object is also wrong. However, this ruler can still distinguish which of two different objects is longer and which is shorter according to the smallest scale------------How can the accuracy remain unchanged despite thermal expansion and contraction?

This post is from Test/Measurement
 
 
 

3232

Posts

0

Resources
7
 
lugl4313820 Published on 2023-4-10 07:14 ADC resolution and resolution 1. How to understand the effective resolution of ADC and the gain error of PGA? 2. What does the resolution of 60000d mean? 3. The resolution of ADC...

8.13ppm/℃---------means that a temperature rise of 1° will result in a shift of 8.13 parts per million?

This post is from Test/Measurement

Comments

In layman's terms, resolution is the maximum scale you can get. Under the reference voltage, 10 bits is 2 times 10, divided into 1024 scales. If your reference voltage is 3.3v, then the ADC resolution is 3.3/1024 = 3.2 millivolts. If your ADC is 16 bits, then it is 3.3/65536 = 50 microvolts. The higher the number of bits, the better the resolution.  Details Published on 2023-4-12 22:52
In layman's terms, resolution is the maximum scale you can get. Under the reference voltage, 10 bits is 2 times 10, divided into 1024 scales. If your reference voltage is 3.3v, then the ADC resolution is 3.3/1024 = 3.2 millivolts. If your ADC is 16 bits, then it is 3.3/65536 = 50 microvolts. The higher the number of bits, the better the resolution.  Details Published on 2023-4-12 22:49
 
 
 

6830

Posts

11

Resources
8
 
QWE4562009 posted on 2023-4-12 18:40 8.13ppm/℃---------meaning that a temperature rise of 1° will result in a shift of 8.13 parts per million?

In layman's terms, resolution is the maximum scale you can get. Under the reference voltage, 10 bits is 2 times 10, divided into 1024 scales. If your reference voltage is 3.3v, the resolution of the ADC is 3.3/1024 = 3.2 millivolts. If your ADC is 16 bits, it is 3.3/65536 = 50 microvolts. The higher the number of bits, the higher the resolution. Under the same sampling rate, the higher the number of bits, the higher the sampling accuracy.

This post is from Test/Measurement

Comments

I understand this.  Details Published on 2023-4-15 12:53
 
 
 

6830

Posts

11

Resources
9
 
QWE4562009 posted on 2023-4-12 18:40 8.13ppm/℃---------meaning that a temperature rise of 1° will result in a shift of 8.13 parts per million?

ADC is a complex thing. Some ADC chips can sell for tens of thousands of yuan, while some ADCs only cost a few cents. A good ADC can produce results with an error of a few microvolts.

This post is from Test/Measurement
 
 
 

12

Posts

1

Resources
10
 

60000d---decimal

This post is from Test/Measurement

Comments

That is 60,000 copies.  Details Published on 2023-4-15 12:54
 
 
 

3232

Posts

0

Resources
11
 
[quote]lugl4313820 posted on 2023-4-12 22:49 In layman's terms, resolution is the maximum scale you can get. Under the reference voltage, 10 bits is 2 times 10, which is divided into 1024 scales. If your reference voltage is 10, the resolution is 10 times the 2 ...

I understand this.

This post is from Test/Measurement
 
 
 

3232

Posts

0

Resources
12
 

That is 60,000 copies.

This post is from Test/Measurement
 
 
 

27

Posts

0

Resources
13
 

Xiaobai came to learn, thanks for sharing, Laotie 6666666

This post is from Test/Measurement
 
 
 

41

Posts

0

Resources
14
 

CS1237 can float up with a breath of output. Normally it can be used as a 16-bit or 17-bit ADC. If it is higher, it needs to be shielded and temperature-controlled.

This post is from Test/Measurement
 
 
 

Guess Your Favourite
Just looking around
Find a datasheet?

EEWorld Datasheet Technical Support

Featured Posts
【Me and gui-guider①】Create a new GUI project

This post was last edited by RCSN on 2021-3-15 23:38 For the effect of porting the PC project done by gui-guider to th ...

How does a MCU software engineer write simple computer software?

I have been engaged in the development of single-chip microcomputers for some years. Thanks to the continuous reduction ...

[Flower carving DIY] Interesting and fun music visualization series of small projects (05) --- WS2812 column jump light

I suddenly had the urge to do a series of topics on sound visualization. This topic is a bit difficult and covers a wide ...

What is the function of the resistor at the power input end of TP4056?

This post was last edited by sky999 on 2022-8-20 03:27 Is the function of this resistor to consume some heat? Looking a ...

[HPM-DIY] openmv for hpm6750 progress 1

This post was last edited by RCSN on 2022-8-26 22:40 When uploading hardware jpeg encoding to openmvide, at 720P (1280* ...

As the industry continues to move toward unified IoT standards

As the industry continues to move toward unified IoT standards, we will continue to support smart home interoperability. ...

[24G millimeter wave radar sensor module review] 24G millimeter wave radar sensor module unboxing and trial use

This time I applied for the HLK-LD116S-24G from Hilink, and was very lucky to be chosen for the only spot available. Fir ...

I made a multifunctional card computer using Allwinner T113, and the cost is only 60 yuan

!(https://bbs.aw-ol.com/assets/uploads/files/1718587521847-qq%E5%9B%BE%E7%89%8720240422143001.jpg) **FunnyPi-T113** is a ...

What is an open-drain output? Does an internal pull-up count as an open-drain output?

Is it considered an open-drain output if the pins of a chip like this are internally pulled up to the power supply? 8181 ...

【Follow me Season 2, Phase 2 Mission】Task Summary Post

This post was last edited by cqutmianma on 2024-10-19 21:51 1. Video display of all tasks The video shows four tasks ...

EEWorld
subscription
account

EEWorld
service
account

Automotive
development
circle

Copyright © 2005-2024 EEWORLD.com.cn, Inc. All rights reserved 京B2-20211791 京ICP备10001474号-1 电信业务审批[2006]字第258号函 京公网安备 11010802033920号
快速回复 返回顶部 Return list