When MAX6675 is connected to a K-type thermocouple probe, the temperature is fairly accurate, about 32 degrees, with a positive or negative error of 2 to 3 degrees. However, as soon as your hand or skin touches the K-type thermocouple wire or probe, the temperature immediately jumps to 10 or 20 degrees. When you take your hand away, it returns to normal.

This post is from PCB Design

Latest reply

https://en.eeworld.com/bbs/forum.php?mod=attachment&aid=NDQxODg0fGMzMTc5OWE3OWU2YTgyOTQ0NGRjNWI5ODlkNzgyNmE2fDE3MzIyODU5NzQ%3D&request=yes&_f=.pdf This is the latest version of the datasheet for your reference   Details Published on 2019-11-3 17:09
 

2w

Posts

0

Resources
2
 

The electromotive force of a thermocouple is only in the mV range, and the voltage of the human body is also several mV. If your hand touches the thermocouple, it will certainly affect the measurement results.

As for the voltage on the human body, you can see it by looking at the electrocardiogram. Each grid on the vertical axis of the electrocardiogram represents 1mV.

This post is from PCB Design

Comments

The K-type thermocouple probe has two wires, one positive and one negative. Touching both of them with my hand will not affect the result. However, if I just touch the iron sheet outside the wire, it will jump around. Why?  Details Published on 2019-8-17 16:53
 
 

768

Posts

0

Resources
3
 
maychang posted on 2019-8-17 16:49 The electromotive force of a thermocouple is only in the mV order of magnitude, and the voltage of the human body is also several mV. If your hand touches the thermocouple, it will certainly affect the measurement results. As for the human body...

The K-type thermocouple probe has two wires, one positive and one negative. Touching both of them with my hand will not affect the result. However, if I just touch the iron sheet outside the wire, it will jump around. Why?

This post is from PCB Design

Comments

I guess the iron sheet outside your thermocouple probe is not well grounded.  Details Published on 2019-8-17 17:06
 
 
 

2w

Posts

0

Resources
4
 
sky999 posted on 2019-8-17 16:53 The K-type thermocouple probe has two wires, one positive and one negative. I touch both of them with my hand without affecting the result. Just touching the iron sheet outside the wire will...

I guess the iron sheet outside your thermocouple probe is not well grounded.

This post is from PCB Design

Comments

Does this iron sheet need to be connected to the ground? I have never seen anyone do this.  Details Published on 2019-8-17 17:23
 
 
 

768

Posts

0

Resources
5
 
maychang posted on 2019-8-17 17:06 I guess the iron sheet outside your thermocouple probe is not well grounded.

Does this iron sheet need to be connected to the ground? I have never seen anyone do this.

This post is from PCB Design

Comments

This question is more complicated. I have not seen your thermoelectric amplifier circuit (MAX6675, including AD conversion, microcontroller, display and power supply), nor have I seen the actual object, so it is difficult to say. I guess your amplifier circuit part (including power supply) is not well grounded, so the amplifier circuit and power supply are not well grounded.  Details Published on 2019-8-17 18:14
 
 
 

2w

Posts

0

Resources
6
 
sky999 posted on 2019-8-17 17:23 Does this iron sheet need to be connected to the ground? I have never seen anyone do this

This question is rather complicated.

It's hard to say without seeing your thermo-electromotive force amplifier circuit (MAX6675, including AD conversion, microcontroller, display and power supply) or the actual object.

I guess your amplifier circuit (including power supply) is not well grounded, so the amplifier circuit, power supply and AC mains will be coupled through the distributed capacitance between the primary and secondary of the transformer (whether it is powered by a power frequency transformer or a switching power supply), and the thermocouple skin is not well grounded. There is a large distributed capacitance between the human body and the AC mains, and the AC mains will be applied to the amplifier circuit through the distributed capacitance between the human body and the AC mains wire and the distributed capacitance between the thermocouple skin and the thermocouple, causing interference.

It is best to connect the amplifier circuit to the ground and the outer skin of the thermocouple to the ground. Or connect the outer skin of the thermocouple to the "ground" of the amplifier circuit. You can try it and see the result.

This post is from PCB Design

Comments

I didn't connect the amplifier circuit, the probe is directly connected to the T+T-  Details Published on 2019-8-17 20:53
 
 
 

768

Posts

0

Resources
7
 
maychang posted on 2019-8-17 18:14 This question is quite complicated. I haven't seen your thermo-electromotive force amplifier circuit (MAX6675, including AD conversion, microcontroller, display and power supply), ...

I didn't connect the amplifier circuit, the probe is directly connected to the T+T-

This post is from PCB Design

Comments

I didn't connect the amplifier circuit. The probe is directly connected to the T+T- of MAX6675. A1 and A2 inside MAX6675 are both amplifiers, and there are analog switches at the input and output of the amplifier.  Details Published on 2019-8-18 16:09
 
 
 

2w

Posts

0

Resources
8
 
sky999 posted on 2019-8-17 20:53 I didn’t connect the amplifier circuit, the probe is directly connected to the T+T- of MAX6675

I didn't connect the amplifier circuit, the probe is directly connected to the T+T-

A1 and A2 inside MAX6675 are both amplifiers, and there are analog switches at the input and output ends of the amplifiers.

This post is from PCB Design

Comments

No, the chip must be connected to GND. Now it can display normal temperature when connected to a 1-meter probe, but it is abnormal when touched by hand. When connected to a 5-meter K-type probe, it is always 0 and no temperature can be collected.  Details Published on 2019-8-19 14:36
No, the chip must be connected to GND. Now it can display normal temperature when connected to a 1-meter probe, but it is abnormal when touched by hand. When connected to a 5-meter K-type probe, it is always 0 and no temperature can be collected.  Details Published on 2019-8-19 14:22
No, the chip must be connected to GND. Now it can display normal temperature when connected to a 1-meter probe, but it is abnormal when touched by hand. When connected to a 5-meter K-type probe, it is always 0 and no temperature can be collected.  Details Published on 2019-8-19 14:21
 
 
 

768

Posts

0

Resources
9
 
maychang posted on 2019-8-18 16:09 I didn't connect the amplifier circuit, but the probe was directly connected to the T+T- of MAX6675. A1 and A2 inside MAX6675 are both amplifiers, and the amplifier input terminal and...

No, the chip must be connected to GND. Now it can display normal temperature when connected to a 1-meter probe, but it is abnormal when touched by hand. When connected to a 5-meter K-type probe, it is always 0 and no temperature can be collected.

This post is from PCB Design
 
 
 

768

Posts

0

Resources
10
 
maychang posted on 2019-8-18 16:09 I didn't connect the amplifier circuit, but the probe was directly connected to the T+T- of MAX6675. A1 and A2 inside MAX6675 are both amplifiers, and the amplifier input terminal and...

By the way, are the two switches automatically turned on when CS is pulled low?

This post is from PCB Design

Comments

"Are those two switches automatically turned on when CS is pulled low?" You need to read the datasheet carefully and analyze its working process to know.  Details Published on 2019-8-19 16:01
 
 
 

768

Posts

0

Resources
11
 
maychang posted on 2019-8-18 16:09 I didn't connect the amplifier circuit, but the probe was directly connected to the T+T- of MAX6675. A1 and A2 inside MAX6675 are both amplifiers, and the amplifier input terminal and...

Should the negative terminal T- of the thermocouple be connected to the external ground or the circuit board to GND?

This post is from PCB Design

Comments

"Should the negative terminal T- of the thermocouple be connected to the external ground, or to the GND on the circuit board?" The thermocouple should of course be connected to the chip pin, and then to the power supply that powers the chip.  Details Published on 2019-8-19 16:03
 
 
 

2w

Posts

0

Resources
12
 
sky999 posted on 2019-8-19 14:22 By the way, are those two switches automatically turned on when CS is pulled low?

"Are those two switches automatically turned on when CS is pulled low?"

You need to read the datasheet carefully and analyze its working process to know it.

This post is from PCB Design
 
 
 

2w

Posts

0

Resources
13
 
sky999 posted on 2019-8-19 14:36 Should the negative terminal T- of the thermocouple be connected to the external ground or the circuit board to GND?

"Should the negative terminal T- of the thermocouple be connected to the external ground or the GND on the circuit board?"

The thermocouple should of course be connected to the chip pins and then to the power supply that powers the chip.

This post is from PCB Design

Comments

Then I am very curious, should the T+T- trace be thicker or thinner? Also, what is the distance from the K-type thermocouple to the circuit board? If copper wire is connected, will it cause errors?  Details Published on 2019-8-22 09:28
 
 
 

768

Posts

0

Resources
14
 
maychang posted on 2019-8-19 16:03 "Should the negative pole T- of the thermocouple be connected to the external ground or the circuit board to GND?" The thermocouple should of course be connected to the chip...

Then I am very curious, should the T+T- trace be thicker or thinner? Also, what is the distance from the K-type thermocouple to the circuit board? If copper wire is connected, will it cause errors?

This post is from PCB Design

Comments

If the distance between the thermocouple and the circuit board is far, a compensation wire should be used in the middle instead of ordinary copper wire. Different types of thermocouples must be equipped with corresponding compensation wires and cannot be mixed.  Details Published on 2019-8-22 09:49
If the distance between the thermocouple and the circuit board is far, a compensation wire should be used in the middle instead of ordinary copper wire. Different types of thermocouples must be equipped with corresponding compensation wires and cannot be mixed.  Details Published on 2019-8-22 09:48
 
 
 

2w

Posts

0

Resources
15
 
sky999 posted on 2019-8-22 09:28 Then I am very curious, should the T+T- trace be thicker or thinner? Also, what is the distance from the K-type thermocouple to the circuit board? If copper wire is connected, will it...

If the distance between the thermocouple and the circuit board is far, a compensation wire should be used in the middle instead of ordinary copper wire. Different types of thermocouples must be equipped with corresponding compensation wires and cannot be mixed.

This post is from PCB Design

Comments

OK, then I feel that the traces in the PCB cannot be too long. After all, they are all copper film lines. Is it necessary for T+T- to be of equal length?  Details Published on 2019-8-22 11:03
 
 
 

2w

Posts

0

Resources
16
 
sky999 posted on 2019-8-22 09:28 Then I am very curious, should the T+T- trace be thicker or thinner? Also, what is the distance from the K-type thermocouple to the circuit board? If copper wire is connected, will it...

“Should the T+T- wiring be thicker or thinner?”

Thickness doesn't matter much, since amplifier input resistance is usually quite high and the current is small.

This post is from PCB Design
 
 
 

768

Posts

0

Resources
17
 
maychang posted on 2019-8-22 09:48 If the distance between the thermocouple and the circuit board is far, a compensation wire should be used in the middle instead of ordinary copper wire. Different types of thermocouples must be used...

OK, then I feel that the traces in the PCB cannot be too long. After all, they are all copper film lines. Is it necessary for T+T- to be of equal length?

This post is from PCB Design

Comments

“Is it necessary for T+T- to be of equal length?” The frequency of thermocouple signals is very low, and the wires on the circuit board do not need to be of equal length.  Details Published on 2019-8-22 11:53
“Is it necessary for T+T- to be of equal length?” The frequency of thermocouple signals is very low, and the wires on the circuit board do not need to be of equal length.  Details Published on 2019-8-22 11:22
 
 
 

2w

Posts

0

Resources
18
 
sky999 posted on 2019-8-22 11:03 OK, then it seems that the traces in the circuit board cannot be too long, after all, they are all copper film lines, so is it necessary for T+T- to be of equal length?

"Does T+T- have to wait that long?"

Thermocouple signals have a very low frequency and it is not necessary for the wires on the circuit board to be of equal length.

This post is from PCB Design
 
 
 

2w

Posts

0

Resources
19
 
sky999 posted on 2019-8-22 11:03 OK, then it seems that the traces in the circuit board cannot be too long, after all, they are all copper film lines, so is it necessary for T+T- to be of equal length?

"Then I think the traces in the circuit board can't be too long, after all, they are all copper foil lines."

You don't understand how thermocouples work. Thermocouples measure the temperature difference, which is the hot end temperature minus the cold end temperature. The hot end is of course the point where the temperature is to be measured, and the cold end is your circuit board. If the temperature point to be measured is far away from your circuit board, the thermocouple must be made very long. However, the two metals of the thermocouple are expensive (especially precious metal thermocouples, such as platinum and rhodium), so the thermocouple cannot be made too long. If the thermocouple is connected to two copper wires, the connection with the copper wire becomes the cold end, and the temperature here is not zero degrees Celsius. Extending from the point where the temperature has dropped, if two metal wires (compensation wires) with the same temperature characteristics as the thermocouple (but cannot withstand such high temperatures) are used to connect the thermocouple, then the cold end is at the other end of the compensation wire, that is, your circuit board. Therefore, the compensation wire cannot be used with the wrong type, and when the correct type of compensation wire is used, the two compensation wires cannot be swapped (the two compensation wires are different alloys).

This post is from PCB Design

Comments

"Then I feel that the traces in the circuit board cannot be too long, after all, they are all copper film wires." The length of the copper wire in the circuit board has nothing to do with the measurement results, provided that the temperature of the connection between the copper wire and the compensation wire is the same as the temperature of the copper wire and the chip pin.  Details Published on 2019-8-22 14:47
"Then I feel that the traces in the circuit board cannot be too long, after all, they are all copper film wires." The length of the copper wire in the circuit board has nothing to do with the measurement results, provided that the temperature of the connection between the copper wire and the compensation wire is the same as the temperature of the copper wire and the chip pin.  Details Published on 2019-8-22 11:55
 
 
 

2w

Posts

0

Resources
20
 
maychang posted on 2019-8-22 11:53 "Then I feel that the traces in the circuit board cannot be too long, after all, they are all copper film wires" You don't understand the working principle of thermocouples. Thermoelectric...

"Then I think the traces in the circuit board can't be too long, after all, they are all copper foil lines."

The length of the copper wire in the circuit board has nothing to do with the measurement results, provided that the temperature at the connection between the copper wire and the compensation wire is the same as the temperature at the connection between the copper wire and the chip pin.

This post is from PCB Design
 
 
 