[Ask if you don't understand] I have never understood the isolation of oscilloscopes, so I would like to ask for help. For example, I accidentally clipped the small clip of the non-isolated oscilloscope probe on the Vcc pin of the 2844 chip in the main circuit of the 220 volt power supply. The other end of the probe was not connected to anything, and smoke was seen on the board. Later, I found out that the oscilloscope was not isolated. How did the smoke come from this situation? My idea is that the 220 volt power supply has a positive pole and a power ground, and the oscilloscope also has its own ground inside. This ground is the potential of the small clip on the probe. What is the isolated oscilloscope usually isolated from?
This post is from Power technology

Latest reply

This post was last edited by iffi123 on 2019-9-8 23:43 topwon posted on 2018-1-20 11:19 By the way, I also changed the schematic diagram of using a 1:1 isolation transformer. The advantage of this is that the oscilloscope case and ground wire can be connected to the ground, which is convenient for the operator...   Details Published on 2019-9-8 23:39
 
 

1374

Posts

2

Resources
2
 
The probe clip of the oscilloscope is connected to the ground of the oscilloscope power cord. Then, if the ground of your board is also connected to the ground of the power cord, the Vcc will be connected to the ground by this clip, burning the power supply.
This post is from Power technology
 
 
 

2w

Posts

0

Resources
3
 
"My idea is that the 220V power supply has a positive pole and a power ground." That's not true. 220V is an alternating current, with only a live wire, a neutral wire, and a ground wire, but no "positive pole".
This post is from Power technology
 
 
 

2w

Posts

0

Resources
4
 
"What does the isolated oscilloscope we usually talk about isolate from what?" It isolates the oscilloscope case (which is also the ground clamp of the oscilloscope probe) from the 220V AC neutral wire (the AC mains neutral wire is connected to the ground at the distribution transformer) and the ground.
This post is from Power technology

Comments

Thanks  Details Published on 2018-1-19 16:37
Thanks  Details Published on 2018-1-19 15:36
 
 
 

1048

Posts

1

Resources
5
 
In fact, it is very simple. Just don't use the oscilloscope with a three-pin power cord with a ground wire. Then be careful not to touch the exposed metal parts on the oscilloscope, such as the ground clamp, the outer ring of the BNC terminal, etc., because these places are floating and may be charged. It depends on where you connect the ground clamp in the circuit, such as the live wire, the neutral wire, or V+.
This post is from Power technology
 
 
 

506

Posts

0

Resources
6
 
In fact, it is recommended that you separate the ground of the oscilloscope during testing and use two power supplies. This will avoid many unexpected events during testing.
This post is from Power technology
 
 
 

1048

Posts

1

Resources
7
 
The safest way is to add an isolation transformer before the AC 220V input of the DUT. This way, you don't have to worry about anything. As long as people don't touch two places with high voltage difference at the same time, they won't be electrocuted. Using an oscilloscope will be no problem. Using a three-pin plug will also be fine. At this time, the point where the ground clip is located is the potential of the earth.
This post is from Power technology
 
 
 

887

Posts

0

Resources
8
 
maychang posted on 2018-1-19 11:52 "What is the isolation oscilloscope that we usually talk about? " It is to isolate the oscilloscope housing (which is also the ground wire clip of the oscilloscope probe) from the 220V AC...
Thank you
This post is from Power technology
 
 
 

887

Posts

0

Resources
9
 
maychang posted on 2018-1-19 11:52 "What is the isolation oscilloscope that we usually talk about?" It is to isolate the oscilloscope housing (which is also the oscilloscope probe ground clip) from the 220V AC...
[1] Since the AC mains neutral line is connected to the earth at the distribution transformer, why do we say that the 220V neutral line should also be isolated from the earth? [2] If it is not isolated, it is easy to connect the GND of the circuit board and the ground of the probe (i.e. the earth), because there is a voltage difference between the two, which will cause a short circuit? Is this understanding correct?
This post is from Power technology

Comments

【1】Since the neutral line of AC mains is connected to the ground at the distribution transformer, why do we also say that the 220V neutral line should be isolated from the ground? I did not say to isolate the 220V neutral line from the ground. What I said on the 4th floor was to isolate the oscilloscope housing (which is also the ground wire clip of the oscilloscope probe) from the ground.  Details Published on 2018-1-19 17:53
【1】Since the neutral line of AC mains is connected to the ground at the distribution transformer, why do we also say that the 220V neutral line should be isolated from the ground? I did not say to isolate the 220V neutral line from the ground. What I said on the 4th floor was to isolate the oscilloscope housing (which is also the ground wire clip of the oscilloscope probe) from the ground.  Details Published on 2018-1-19 17:50
 
 
 

2w

Posts

0

Resources
10
 
shaorc posted on 2018-1-19 16:37 [1] Since the neutral line of AC mains is connected to the earth at the distribution transformer, why do we still say that the 220V neutral line should be isolated from the earth? [2] If not...
[1] Since the neutral line of AC mains is connected to the earth at the distribution transformer, why do we still say that the 220V neutral line should be isolated from the earth? I did not say to isolate the 220V neutral line from the earth. What I said on the 4th floor was to isolate the oscilloscope housing (which is also the oscilloscope probe ground clip) from the earth.
This post is from Power technology
 
 
 

2w

Posts

0

Resources
11
 
shaorc posted on 2018-1-19 16:37 【1】Since the neutral line of the AC mains is connected to the earth at the distribution transformer, why do we also say that the 220V neutral line should be isolated from the earth? 【2】If not...
【2】If not isolated, it is easy to connect the GND of the circuit board and the ground of the probe (i.e. the earth) because there is a voltage difference between the two, which will cause a short circuit? Is this correct? It is not "easy". If not isolated, as long as there is a voltage between any point on the circuit board and the earth (whether AC or DC), it will definitely cause a short circuit through the oscilloscope probe ground clip (connected to the earth).
This post is from Power technology

Comments

Thank you, thank you  Details Published on 2018-1-20 11:05
 
 
 

525

Posts

235

Resources
12
 
It is recommended that the OP first understand the principles of TT/TN-C/TN-CS in the power supply system.

This post is from Power technology
Personal signature爱电子,爱生活
 
 
 

1048

Posts

1

Resources
13
 
I made a simple schematic diagram for easy understanding. If the ground pin in the oscilloscope's power cord is connected to the earth (the hexagonal star), then when the probe's ground clip is connected to a place with high voltage relative to the earth, such as the live wire in the figure, the live wire high voltage will be short-circuited to the earth by the oscilloscope's ground clip, which will burn the components on the circuit. A relatively simple solution is to disconnect the hexagonal star. I will make a special power cord, remove the middle ground pin on the plug, and replace the oscilloscope with this power cord when testing the primary circuit of the isolated power supply or the non-isolated circuit. In addition, sometimes when it is necessary to measure the voltage drop across the resistor, this power cord will also be used when the ground clip is not at ground potential.

示波器地线隔离.png (27.89 KB, downloads: 1)

示意图

示意图
This post is from Power technology

Comments

Thank you so much, I just need this picture to understand, thank you thank you  Details Published on 2018-1-20 11:04
 
 
 

887

Posts

0

Resources
14
 
topwon posted on 2018-1-20 09:41 I made a simple schematic diagram for easy understanding. If the ground pin in the oscilloscope's power cord is connected to the earth (where the hexagon is), then when the ground wire of the probe...
Thank you so much, I need this diagram to understand, thank you thank you
This post is from Power technology
 
 
 

887

Posts

0

Resources
15
 
maychang posted on 2018-1-19 17:53 【2】If it is not isolated, it is easy to connect the GND of the circuit board and the ground of the probe (i.e. the earth), because there is a voltage difference between the two, which will cause a short circuit? Is this understanding correct...
Thank you, thank you
This post is from Power technology
 
 
 

1048

Posts

1

Resources
16
 
By the way, I also changed the schematic diagram to use a 1:1 isolation transformer. The advantage of this is that the oscilloscope casing and ground wire can be connected to the ground, which guarantees the safety of the operator. The disadvantage is that the isolation transformer is more expensive and takes up space (the specific choice is related to the power of the product).

示波器地线隔离2.png (30.15 KB, downloads: 0)

加隔离变压器

加隔离变压器
This post is from Power technology
 
 
 

3980

Posts

0

Resources
17
 
The most economical way: use pliers to twist off the PE on the three-pin plug of the oscilloscope, turning it into two pins (only L and N) The most advanced way: buy an isolation probe
This post is from Power technology
Personal signature

YesWatt艺瓦特电子科技有限公司 傻大粗电源转换器制造商 https://apu5ob0ydv0ysskfm03hs4dtqfr97j68.taobao.com/

 
 
 

2

Posts

0

Resources
18
 
This post was last edited by iffi123 on 2019-9-8 23:43
topwon posted on 2018-1-20 11:19 By the way, I also changed the schematic diagram of using a 1:1 isolation transformer. The advantage of this is that the oscilloscope case and ground wire can be connected to the ground, which is convenient for the operator...